ResMed AirCurve 10 VAuto Setup Bilevel : Clinical Menu Set up and Settings Explained

  Рет қаралды 115,256

TheLankyLefty27

TheLankyLefty27

Күн бұрын

Пікірлер
@DennyBob521
@DennyBob521 10 ай бұрын
I’ve been on BiPAP since January 2006 and I believe it saved my life. I’ve been adjusting my pressure up and down to keep my AHI as low as possible for years, with a Respironics Dreamstation, and a System One before that. So I’m new to ResMed Vcurve Auto 10. Thanks for the info, I find it unusual that my IPAP and EPAP are so close together based on what I’ve seen here, and what my friends have for pressures. Mine are 10 and 12. That’s what I’ve been prescribed. I’ve adjusted those over time. I’ve lost 70 pounds since April 2023 and 100 total since Sept 2021, so I really need to change my pressure. I think it’s too high for a much skinnier me.
@BabyteethMore
@BabyteethMore 10 ай бұрын
Same! Mine were prescribed very high but only a difference of 2! I am adjusting both down in 10% increments due to morning ribcage soreness the high pressure is causing. So far so good.
@michaelvercimak2594
@michaelvercimak2594 2 жыл бұрын
When I was prescribed BiPAP my A:H index was 46 and I was desaturating down to 72% with arrhythmias.. When I went to get help from a home supplier they told me no machines were available and it would be six months at least before I could get one. Several suppliers were rude an in the end they left me twisting in the breeze. I bought an air curve 10 off the internet but could not get in the menu to set the pressures. This video showed me how and probably saved my life.
@the-mentat
@the-mentat 2 ай бұрын
Very insightful. I just switched from CPAP to BiPAP. My first night, I felt the inhalation stopped prematurely, which was annoying. This video helped me understand the settings properly. I actually did change the Ti Max from the default of 0.3 to 2.0 to 1.0 to 4.0. I tested my average inhalation time, at least while being awake, and it was 2.5 to 3.5 seconds. I noticed that even then it was too sensitive so I adjusted the cycle setting to low and it became much easier and normal feeling breathing.
@cathysilva5011
@cathysilva5011 3 ай бұрын
I’m lost I have to watch this 100 times over
@cyberfiche
@cyberfiche 3 ай бұрын
Best video covering the clinical menu! I've been trying to figure this stuff out for a couple months. This explains so much!
@NathanKrueger-z9k
@NathanKrueger-z9k 8 ай бұрын
Your a true hero, you just saved hundreds of dollars in bs doc visits 2 times per month @ $300 a pop!! Keep up the good info, God bless!! Thank you!!!
@ellenbrown5542
@ellenbrown5542 2 жыл бұрын
I started my bipap therapy a couple years ago. I have used it on and off. I started pulling my mask off during the night so I quit using it all together. Cardiac Dr said it will worsen my afib if I don’t use it, so I went back to the lung Dr. That was a year ago. He said he would call and have my settings changed. After 6 phone calls to the Dr’s office and several to Lincare it’s still not adjusted. It will be another month or more to get back into the Drs office to get the settings changed. I am going to use this video to try to set it myself. Thank you for the help.
@willkendallpro
@willkendallpro 2 жыл бұрын
This video is SO GOOD, I watched the ad.
@ohplezz
@ohplezz 3 жыл бұрын
I own this machine and it is fantastic. Without this machine, I would have given up on cpap therapy.
@robertlongoria3748
@robertlongoria3748 3 жыл бұрын
Thank you soo much. I had given up on my machine because I was having trouble breathing . After watching this video I was able to adjust the settings to my liking and not what the doctor recommended which obstructed my breathing. Now I can breathe and get a good nights rest. I appreciate what you do. Thanks again
@cosmicagape7
@cosmicagape7 3 жыл бұрын
My aircurve10 vauto doesn't have a option for me to adjust pressures. Ineed help finding the way to adjust
@jackdaniels423
@jackdaniels423 3 жыл бұрын
@@cosmicagape7 all of them do. Hold your dial and middle button down for 3 seconds
@jeremyoboe
@jeremyoboe Жыл бұрын
Best video on the Internet about Bipap machine settings. Thank you!
@wetworkwolf
@wetworkwolf 22 күн бұрын
Thank you so much! I was snoring very lightly while wearing the cpap and driving my spouse crazy. With the info I should be able to fix the issue.
@breathlessinbedvlogs
@breathlessinbedvlogs 3 жыл бұрын
thank you for doing the lord's work out here and decoding these machines 🙏🙏
@DC-8-8-8-
@DC-8-8-8- Жыл бұрын
🥰😁😆 yesssssss so much 😇🙏
@underthebus5801
@underthebus5801 3 жыл бұрын
I use an AirCurve10 Vauto. I didn't realize how complex it is. My doctor is pleased with my results thus far so I don't touch the settings except adjusting the time during the time change in Spring and Fall.
@brothergrok
@brothergrok 3 жыл бұрын
YEAH IT's my machine! Let's do this, Jason! *listens intently*
@Freecpapadvice
@Freecpapadvice 3 жыл бұрын
Dang! I JUST hit submit. Griff doesn't play!
@Gabe73C
@Gabe73C Жыл бұрын
Thank you. I've lost a 100lbs since my sleep study and trying to get pressure reductions have been three ring circuses. 👍
@JLone55
@JLone55 Жыл бұрын
Congrats to you. I'm trying to follow your footsteps. Very difficult for me
@Gabe73C
@Gabe73C Жыл бұрын
@@JLone55 It takes time. Best thing to do is find a hobby or sport that gets you moving and that you actually enjoy doing. For me, it was riding a bike. Now I mountain bike regularly because I want to. Good luck 👍.
@Chahtawarrior31
@Chahtawarrior31 2 жыл бұрын
Thank you, Got my Aircurve sent wide open and had two of the most sleepless nights. Watched your video dialing in my IPAP 17cm/EPAP 12cm with PS and walla! Semper Fi OORAH
@catallergiccat
@catallergiccat 7 ай бұрын
One important difference between s-mode and vauto is that s-mode doesn’t records flow limitation if you are using Oscar to inspect your data.
@irvinggreines822
@irvinggreines822 2 жыл бұрын
Congrats: You have boggled my mind.
@Moveon10001
@Moveon10001 2 жыл бұрын
My RT refused to tell me how to get into the machine extra settings. I figured it out anyway. I just had another sleep study because I was having over 40 episodes an hour using my new ResMed Air Curve 10 VAuto machine. I used the smartphone app to tell me the data. I have central sleep apnea so they did another sleep study to figure out my settings. Thanks for all the information about the pressure support etc. That is very useful. I was wondering what the difference in V auto was from the S therapy settings and CPAP.
@chuckmitchell2713
@chuckmitchell2713 2 жыл бұрын
I'm so glad I found your channel and these great videos (even with the corny jokes! ;-)) I've fought with sleep apnea for over 15 years, through several incompetent DMEs and doctors, and, of course, the recall. When I received my AirCurve 10 six weeks ago I was so hopeful that it might finally solve my issues. Instead, I had horrible numbers with dozens of rapidly repeating CA hits. My current DME and doctor were unable, or unwilling to read the charts. So with your help I figured them out and with this video have been learning to adjust the parameters to finally get real sleep the past few nights. Thank you. Thank you. Thank you. I'm checking your other webpage to look into having a consultation in the near future.
@vinnyg2619
@vinnyg2619 2 жыл бұрын
Thanks for the video! I had to replace my Aircurve 10 S due to the motor whirring and the DME replaced it and gave me a Vauto, so you gave me a quick lesson on how to set it with auto. If it doesn't work out I'll set it to "S" mode and use it as a standard bilevel machine. Thanks again!
@azjack827
@azjack827 2 жыл бұрын
Wow i have watched this video 5+ times. I sorta or kinda understand but I’m not going to mess with the settings except for auto start and ramp. Thanks so much 😊
@ralphb4012
@ralphb4012 2 жыл бұрын
Relative humidity is the % of moisture in the air compared to the total amount of moisture it can hold before the air cannot hold moisture before it "rains" for simplistic terms. Colder air holds less moisture than warm air.
@faanzeonlam3219
@faanzeonlam3219 Жыл бұрын
Thank you so much, I was literally drinking dripping water in middle of the night before lollllllll
@markdeegan7268
@markdeegan7268 Жыл бұрын
A great down to earth presentation
@af1966
@af1966 Ай бұрын
Coming from a AirSense 11 AutoSet using AutoSet (APAP) mode, I use EPR on of 3 full-time, AutoSet Response of soft. . No ramp. Min Pressure 12.2, Max pressure of 14.4. After watching this video I am not sure what parameters I should set and to what values for the Aircurve 10 using VAuto mode. These setting work well for me as I view my nightly data in OSCAR. I am doing this on my own so I don't have a medical professional to consult. Could you offer some help? Thank you.
@brandonasarisi558
@brandonasarisi558 2 жыл бұрын
I've been struggling. Needing to up my pressure for a year now and I just found this video. Thank you so much.
@bigkeoni6429
@bigkeoni6429 3 жыл бұрын
I have EPAP set to 4.0 and IPAP to 12.0 - Is that too big of a variance. It seems to work. I can average 4 - 5 hours on the machine with really good nasal pillow seal and less than 4 apnea events/hr. I was diagnosed with 50 events/hr
@wacopiper
@wacopiper Жыл бұрын
Why won't techs tell us this stuff? Thank you Lanky Lefty!
@gardengrower7633
@gardengrower7633 Жыл бұрын
They want to keep their jobs and view us patients as incompetent and not trustworthy/reckless concerning our personal health decisions. Not all, but most.
@geraldchilders2625
@geraldchilders2625 11 ай бұрын
I like to control my settings makes things much easier only you really no exactly how you feel if you need more pressure or less been living with it my hole life and after I got my first machine and started messing with it I don’t really need to go back to doctor very often every few years to get new machines that’s it
@AndrewHedlund100
@AndrewHedlund100 2 жыл бұрын
Thank you for making this video I have yet to find my rhythm with CPAP after being diagnosed over a year ago and my doctor is kind of flaky. I find myself consciously breathing a lot even when awake and when I stop thinking about my breathing and my breath softens there’s a hell of a lot of resistance on the Exhale most in my nasal area So I’m trying to find the right pressure on the exhale I don’t think the inhale this is much of a problem thanks
@UUBrahman
@UUBrahman 2 жыл бұрын
Try a difference of 2, so inhale is set to 10, exhale might be set to 8. Just a suggestion.
@darkandlovely
@darkandlovely 26 күн бұрын
I didn't see this in the video, and at my facility we use these machines all of the time, what happens when you turn PS off in VAUTO mode? Does it mimic "S" mode then or is it auto CPAP? Is there an Auto CPAP on this machine or a way to mimic it? TIA
@amorphousblob
@amorphousblob 3 ай бұрын
Am I correct to assume the bilevel algorithm doesn't have the same problems as EPR, where it often ramps up too late during the breath to be useful (thus having to compensate by raising the min pressure 1-3 depending on what EPR is set to)?
@ericdavis1964
@ericdavis1964 5 ай бұрын
Ty, my wife has been struggling but the doc says the numbers look good so she dosnt want to change it with 10 ahi an hour where the standard is 5 and under i will give her a week them use this video to do it on our own while we look for new doctor that will take numbers and what the person is feeling
@jerryparz4103
@jerryparz4103 2 жыл бұрын
I'm new with this BIPAP. I have settings IPAP 19 and EPAP 15. My events in the past month have only been above 2, 2 times. Most are 1 or less. Can I just reduce the 19/15 down? Like 18/14 etc. until events are between 3 and 4 per hour? Wouldn't reducing these two settings be better for my comfort?
@randallfraley2810
@randallfraley2810 3 жыл бұрын
Could you do a review of the setting on the Air curve10 ST machine like you did on this machine?
@EmmaDee
@EmmaDee 2 жыл бұрын
Mine has an S after the curve too. Air curve 10S, so your is an ST?
@thrillbilly6147
@thrillbilly6147 2 жыл бұрын
Recovering from Turbinate reduction surgery. My original pressure setting is too much. Thank you for the instructional video on pressure settings.
@sambaggins2798
@sambaggins2798 3 ай бұрын
Thank you. Easy explanation with an easy for to follow tutorial. It helped me a great deal.
@timstoytalk
@timstoytalk Жыл бұрын
My dr said there is an adjustment for altitude but I forgot. Can you please help with this adjustment?
@keithiekel3274
@keithiekel3274 Жыл бұрын
Question, but first Thank you for doing this you might have just saved my life. I will try and be as brief as possible. I have this bipap but my sleep therapy session was with the Phillips dream and we just determined my iPap and epap which was 17 over 11. And it worked great for me my lungs are bad born with asthma and COPD in 2000. I am over 300 pounds. I was on two different recalled Phillips units for over two years both had factory recalls when given to me. Now my lungs are less then 38% i had to get this respironics talking medicaid into paying for a third one as Phillips never replaced the recalled ones. Its hard for me to leave the house so my wife picked this model up and they just gave me the same pre set numbers. But my condition had worsened since I first went on a bipap over two years ago. So my problem I'm having is , one I feel like I am not getting enough pressure and two that my breath when breathing in it isn't picking up and shoving the air in fast enough and it's not keeping it going in long enough. The meter I watch climbs and maxes out and I am still breathing in. So I changed my iPap after starting at 19 came down to 18 that felt right but kept my epap at 11. But I am having the same problem. Should I go back to original numbers 17 over 11 and raise my trigger time to high ? Please help me as I am terrified of falling asleep as I am waking up with bipap after ten mins having 7 times stopped breathing. I'm exhausted and don't trust anyone in this industry right now after Phillips and the medical supply company giving me a federally recalled unit two different times a year after recall and let me use them for two years. Thank you sorry it's long I'm tired.
@geraldchilders2625
@geraldchilders2625 11 ай бұрын
I felt like I wasn’t getting enough air I tried cpap level 20 and I get good sleep now I could never find the right setting on vpap mode I tried lots just never felt right
@matty00926
@matty00926 3 жыл бұрын
I think you have the rise time backwards, 900ms is a much slower rise from EPAP to set IPAP, than 300, 200, min, etc,. Rise time at min is essentially a whack in the face.
@Freecpapadvice
@Freecpapadvice 3 жыл бұрын
Is that not what I said? 100 milliseconds is faster than 900 milliseconds. Time it takes to reach full inhalation is 100 milliseconds, faster than taking 900 milliseconds to reach full inhalation. If I didn't say that clearly I'll have to pin this comment! If I said it incorrectly, thanks for catching it Matthew! But if you're wrong....then SHUT UP MATTHEW!!!! 😂😉
@matty00926
@matty00926 3 жыл бұрын
@@Freecpapadvice kzbin.info/www/bejne/qqScnYGAZ8xkf6c This is the part of your video I was referencing. Lol either way I still love you
@blackrockcity
@blackrockcity 3 жыл бұрын
@@Freecpapadvice Seems like Matthew is right.
@gardengrower7633
@gardengrower7633 Жыл бұрын
@@Freecpapadvice yes you did say it incorrectly. I was caught off guard too. :)
@dfunkmale
@dfunkmale Жыл бұрын
I wish this machine had a "fall time" setting. Adjusting the rise time prevents me from feeling like I'm getting smacked in the face. But since the switch back to EPAP is instantaneous, it makes me feel like the rug is getting pulled out from under me. That's why I use easy breathe, even though I suspect it prevents me from spending enough time at my prescribed ipap pressure
@papajim1974
@papajim1974 2 ай бұрын
How & where do you attach a bacterial filter in Aircurve 10S?
@redhotchillipepper9917
@redhotchillipepper9917 Жыл бұрын
Min is 6-8 cms of water in order to overcome the upper airway resistance. I would suggest Epap of 5 for Ipap of 11 as a starting point and take it from there if your using an Vauto
@jimbauer5560
@jimbauer5560 3 ай бұрын
Hopefully you will see this, my vauto was set at 8/22 with a ps of 4….should my ps be increased?
@calvinmay8726
@calvinmay8726 Жыл бұрын
When using the humidifier how do you keep it in chill mode. My aircurve will always go back to heated . When i first started using the machine there was an option to switich between chill / heat. I dont use heated tube.
@ThumperSillywabbit
@ThumperSillywabbit Жыл бұрын
Jason. Could you please go over once again the pressure support issue. Are you saying we should just use the s mode as v mode is so flawed and there is the issue of encouraging central apneas. Hellllp.
@randytaylor7828
@randytaylor7828 2 жыл бұрын
New to the Bpap world. Great video. I had two sleep studies. Then a machine was sent to me. I'm having the same issue at home as I did during my sleep studies. Everything feels great until my pressures ramp up so high it compromises my face mask. This wakes me up. I complained about this during the study but the machine I received does the same thing. I can't tighten my mask any more because it's uncomfortable.
@heavydownpour560
@heavydownpour560 2 жыл бұрын
I have the same problem. My iPad is set at 21.0 and my epiphany is at 17.0. When it ramps up and this is to its fullest potential, I have leaks on the side and in the eyes. I’ve tried the dreamwear that fits under the nose and over the mouth but I don’t sleep as well with it. It feels better on my head but doesn’t work as good as my full face mask with the straps that I adjust and readjust about every night. I’ve made myself wear it and I went from 76 events to 3 and I still feel rested even with 5.5 hours avg sleep. I have a beard and I know it causes some leakages. Best thing to do is talk to your dr or the folks that sent you your machine and get them to help you with it. I don’t know if I’ve been any help, but I can sympathize with you.
@Neighvmom
@Neighvmom Жыл бұрын
Randy my husband has the exact issue you do. We set it at 45 minute ramp, at the end of the ramp it kicks the pressure up so high, it blows the mask off his face and wakes him up. Did they find a solution for you? We have had this machine a month and they can't seem to make any more adjustments to see if he can stop the pressure from jumping so high at the end of ramp. In his sleep study, the machine did the same thing, so the tech put a rubber band thing on my husbands head to keep the mask on his face. Frustrated
@donnaclifford3688
@donnaclifford3688 Жыл бұрын
Try asking your doc if you can turn off the ramp time and go by your set pressure i did that and i have very little issues
@spamcan9208
@spamcan9208 11 ай бұрын
Your pressure sounds too high. Ramp time is a lie, there's no ramping. My "ramp" pressure nearly doubled instantly after the timer expired. The pressure was so intense that it was no different than sticking your head out of a car window on the highway. My cheeks were fully puffed out like Dizzy Gillespie.
@BlockChainzz
@BlockChainzz 11 ай бұрын
I was having the same issues and bought the pillow mask that just goes on the nose, surprisingly I Can sleep with my mouth shut haha seems to work better and it’s easier to yawn ect. but am still not wearing it through the night. Going to try and play with the ramping. Hence why I’m here…
@cynthiaybarra7317
@cynthiaybarra7317 2 жыл бұрын
If I'm understanding this correctly, my prescribed EPAP of 15 is what is controlling my obstructive apneas? My IPAP of 19 is what is handling my centrals and hypopneas? I'm on bilevel because I couldn't tolerate CPAP during my sleep study 5 years ago. I struggled exhaling against the pressure and didn't fully fall asleep until the tech switched me to bilevel. I recently lowered my pressures by 1 and I've noticed my high leak rate has drastically reduced, my centrals have reduced, my osa hasn't worsened, and I've never had many hypopneas. I'm thinking about reducing my pressure relief to see if I can get it to where, in the future, I can get a CPAP machine instead of a bilevel, but I thought the IPAP would be what handles my osa, not my EPAP, that's very interesting. I'm excited to try reducing my IPAP, it would be AMAZING to have my pressure be 14!
@MrAntonNYC
@MrAntonNYC Жыл бұрын
Hey lanky I had my aircurve for about 2 years now. I lost 50 pounds and my iPap is 22.0 and my epap is 18. My chest feels sore I think my pressure is too high I average up to 2 events a night should I switch my settings
@Freecpapadvice
@Freecpapadvice Жыл бұрын
If you lost 50 lbs, you definitely need to reevaluate your pressure.
@MrAntonNYC
@MrAntonNYC Жыл бұрын
@@Freecpapadvice can I do it threw your app
@Freecpapadvice
@Freecpapadvice Жыл бұрын
@@MrAntonNYC you can use OSCAR to evaluate your breathing with bilevel and decrease until it stabilizes.
@AineDerwin-ng2hc
@AineDerwin-ng2hc Жыл бұрын
Hey, Can you please explain apnoea alarm on an Aircurve 10. Thanks in advance
@samsonlo7058
@samsonlo7058 2 ай бұрын
Hi lankylefty I have a question ;my current place of employment is saying that in VAuto mode if you leave the pressure support at 0 that makes the VAuto into a autoset cpap. I’m confused as my reading the manual doesn’t support that thought. I could be wrong but I thought APAP is Resmed’s mode for that. I can’t get anyone on Resmed to call me back so wondering what your take is? If the doctor orders autoset with a max epap of 20 and a min epap 5 yet I place him on VAuto with ipap 20 and epap 5 and zero PS isn’t that the same?
@samsonlo7058
@samsonlo7058 2 ай бұрын
Again we are using the Aircurve 10 VAuto.
@ThumperSillywabbit
@ThumperSillywabbit Жыл бұрын
I loved your advertisement. 😊
@k8eekatt
@k8eekatt 3 жыл бұрын
Thank you for making information accessible
@fletcher3913
@fletcher3913 3 жыл бұрын
Thank you, Jason. I'll have to get my machine and watch this again.
@JamesHobart-k1n
@JamesHobart-k1n 2 ай бұрын
Your explanation has been very helpful but I can’t figure out how to extend my exhalation time. I feel like the machine is forcing me to breathe very fast.
@Freecpapadvice
@Freecpapadvice 2 ай бұрын
Set Trigger to low or very low.
@donnaclifford3688
@donnaclifford3688 Жыл бұрын
Hey lanky lefty i have a resmed air curve 10 i just checked the settings you just showed on your channel and i checked them and it showed that i have slim line tubing how can i tell the difference between climate line and slim line tubing?
@latwood101252
@latwood101252 Жыл бұрын
I am using the AirCurve10 Vauto. Does this unit have Bluetooth connectivity? Cell phone service is very weak to not at all available.
@bleednate
@bleednate 2 жыл бұрын
Does SmartStart also supposed to automatically turn off the machine when the mask is removed? My AirSense 10 has this feature and it works for both auto-on and auto-off, but my auto-off doesn’t seem to work with my AirCurve 10 Vauto.
@brooklynqueensvblizzel257
@brooklynqueensvblizzel257 Жыл бұрын
Hello I wake up with a very dry throat what can I do for that Thank You.
@nonstar8182
@nonstar8182 5 ай бұрын
My DME told me that the Leak Alert (audible) is not on the VAUTO.. only the S/ST. Is this true? I see your menu shows it.
@PuggleLover101
@PuggleLover101 2 жыл бұрын
Great video, i learned a lot. One thing I'd be interested in is to see an example script and to understand exactly what variables the doctor will prescribe and what variables are expected to be set by the patient
@gwkdad
@gwkdad Жыл бұрын
I just picked up a loaner aircurve and it doesn't display my AHI's on the screen when I select Sleep Report, how do I add that variable to the screen? Thanks.
@l.dcrouse7150
@l.dcrouse7150 2 жыл бұрын
I have changed the settings on the moisture also and try using it without any..maybe lower the temperature on the tube??if I turn it up it drowns me with condensation
@caroltinnea4922
@caroltinnea4922 Жыл бұрын
I use this machine. I still wake up because I have severe dry mouth and/or indigestion and feel bloaty. How can I adjust the settings to prevent this?
@Laila_21k
@Laila_21k 2 жыл бұрын
Thank you for the video. My mother has this machine and it doesn’t stay on all night and she is currently not getting enough sleep. Is there a way of checking that?
@sharonchriswell8500
@sharonchriswell8500 3 жыл бұрын
This is very helpful. We cant get a sleep test till July and my husband cardiologist said his sleep apnea is life threatening. We got the air curve on our own, but know nothing. Your video has helped with my husbands sleep. We would like to do a zoom even if there is a fee. How do we do that?
@hammettgirl
@hammettgirl Жыл бұрын
Me too! Did you get a Zoom?
@kalelklenk
@kalelklenk Жыл бұрын
So the rise time comes to do what the ST mode does??..
@samanthastevens4197
@samanthastevens4197 Жыл бұрын
Great information - Much needed - Thank You.
@JimCorso-fk9ro
@JimCorso-fk9ro 7 ай бұрын
Please tell me what battery I should buy for my ResMed Air curve 10
@UsernameUnkwn21
@UsernameUnkwn21 2 ай бұрын
Ask Resmed
@keicol
@keicol Жыл бұрын
Do have a video on the Aircurve 10 ASV ? Mine has different settings than this Vauto, I get all your settings for the Vauto, nicely explained, I wish I had those same options. Any suggestions?
@Mary-xf4ew
@Mary-xf4ew Жыл бұрын
Can I get a portable oxygen concentrator that will be compatible with a Dream station 2 cpap machine
@Tonimadroni
@Tonimadroni 7 ай бұрын
Is S mode the same as fixed bipap essentially?
@therandomperson6354
@therandomperson6354 Жыл бұрын
Great video thanks I Currently on cpap but pressure's are hight every night 18-20 apneans in teens so looking like this bipap what would you set it too pressure wise?
@geraldchilders2625
@geraldchilders2625 11 ай бұрын
I set mine to 25 -20
@scottgreear4591
@scottgreear4591 Жыл бұрын
Wow!!!! I have a cpap and I am desating. My Mac preasure av preasure is 11 (5/15) and my ahi is .5. What should my settings be. 88 or less 15 minutes. 347 desat events. Change settings to 10/15 and ahi stays 1/2%. Or .5 and desat stays same but 88 or less goes to 5 minutes
@elstranger98
@elstranger98 Жыл бұрын
Just found this channel and I am new to this bipap. I have been struggling with the mask. I tried the full mask and now the F30i. They always have air leakage and it seems I have constantly tighten the mask, which makes me uncomfortable. I checked the settings and show Mode Vauto Max Ipap 18, Min Epap 13, and PS 3. Do you any suggestions if maybe the settings can be adjusted or a better way to deal with the mask?
@catallergiccat
@catallergiccat 2 жыл бұрын
Recently I switched to BiPAP, and found lots of the CA (clear airway) events (28). I don’t know the number when I was using CPAP, but the overall AHI increased from around 3 to around 5. My current setup is 4-13 with PS 3, I wonder if that 3 causes more central sleep apnea events. Is it worth adjusting the number a little higher? Or should I get back to CPAP?
@honorwashpressurewashingll9056
@honorwashpressurewashingll9056 3 жыл бұрын
Hi Jason...my doctor just increased my pressure but now it's forcing air in when I need about another second of exhale. Which setting do I need to adjust? Thanks!
@Freecpapadvice
@Freecpapadvice 3 жыл бұрын
Decrease Trigger sensitivity, increase Cycle sensitivity. See if that helps first.
@honorwashpressurewashingll9056
@honorwashpressurewashingll9056 3 жыл бұрын
Hey Jason, I tried decreasing trigger sensitivity and increasing cycle sensitivity but still didn't help. Any other ideas to lengthen the inhale and exhale? Thanks for your help!
@CChen-wd9kg
@CChen-wd9kg 2 жыл бұрын
Thanks for the explaining; one question is the Pressue Support(PS) level the same as chosen even during Ramp time? If so, if PS is 5 and starting during ramp(assuming starting EPAP=4), IPAP= 9, is high IPAP a concern if you want to maximize the comfort and tolerability of using the machine.
@abhishekjuneja9869
@abhishekjuneja9869 3 жыл бұрын
Thank you for your video Jason. I have recently bought a Resmed Aircurve st model. How is it different from vauto.Also I tried connecting it to wifi or phone network but couldn't.Can you help me out?
@dianawitt1197
@dianawitt1197 4 ай бұрын
It has its own WiFi built into it..
@jmangames5562
@jmangames5562 Жыл бұрын
Well I find I breathe slow long breaths while sleeping and the machine on vauto cuts my breath off, I am still inhaling and it stops. This wakes me up every single breath doing this. It seems rather stupid to me that is vauto and would cut your breath off?? how do I get it to stop because if I can't I will never be able to use it.
@lorenzoishere
@lorenzoishere 2 жыл бұрын
Hi, can you make a similar video with the AirCurve ASV please? dont fully understand the difference beatween Bilevel auto and ASV. I´ve been recomended with one of those to improve the UARS. Thanks
@geraldchilders2625
@geraldchilders2625 11 ай бұрын
So how do you no what’s a good ps setting
@robertbrummerstedt4345
@robertbrummerstedt4345 8 ай бұрын
Great information for AirCurve 10 VAuto Thanks
@charlespaine987
@charlespaine987 Жыл бұрын
Question my new machine does not show “easy-breathe setting . Am I missing another setting or did new machine drop that setting
@Freecpapadvice
@Freecpapadvice Жыл бұрын
It does seem that the newer machines are dropping that setting.
@jerrydenomme4485
@jerrydenomme4485 3 жыл бұрын
excellent video I do have a pulmonologist that I can't get to send in a prescription change and this video at least helps me figure out how to experiment. I set up the bipap with pressure support of 5 with min and max 6 & 12. My air app gave me 3.2 events not sure what i should do next. Should I increase pressure support to 6 and keep the min max the same or also increase the max to lets say 14 any help would be great thanks
@1mimidog
@1mimidog 3 жыл бұрын
Just got my new Resmed Curve. From first night my apneas increased. My old apap was set at 10.9 this bipap is at 9 Dont think its high enough, waking up with headaches and nsusea.
@stevepfefferle7095
@stevepfefferle7095 Жыл бұрын
My bipap just went bad about a week ago. They said my machine was too old to fix( 5 ys old). The new one cost me 1700. Not even sure if my insurance will cover any of the cost. Just received it 3 days ago. They said the bipap is the most expensive. Is this true?
@michellelippincott6097
@michellelippincott6097 3 жыл бұрын
Have a question, have you seen improvement in patients for REM sleep disorders using cpap therapy with combination of the apnea, hypopneas and severe fragmentation sleep disorder symptoms ??? ~Michelle
@drachenfeuer5042
@drachenfeuer5042 Жыл бұрын
Mine ran for awhile my first night and shut off what’s going on!
@gigilovesbecca
@gigilovesbecca 3 жыл бұрын
This is an odd question, but is there a way to turn the alarm volume up?
@UUBrahman
@UUBrahman 3 жыл бұрын
I have the same question on this model of machine. It is not very loud.
@SnarleySteve
@SnarleySteve Жыл бұрын
I have both obstructive and central apnea. I was recently switched to this machine because the cpap didnt take care of the central. I tend to take long slow breaths when I sleep and the machine increases the pressure before i'm done exhaling. Is there a way to control this.
@dfunkmale
@dfunkmale Жыл бұрын
I have the opposite issue. I've found adjusting the trigger and cycle settings to be helpful. In your case, what you might want to try is decrease the trigger sensitivity. If you're a slow, shallow breather, maybe that will prevent some "false" ipap events. You can also consider increasing the cycle sensitivity, which will make it easier for you to initiate switching to the epap pressure.
@jvkrause
@jvkrause 3 жыл бұрын
Thanks for this video and the info about the special sauce of bi-level!
@jimquantic
@jimquantic 2 жыл бұрын
Thanks but I don't have these menu options at all on my ResMed10 machine, don't say any way to access them either.
@garypeterson3628
@garypeterson3628 2 жыл бұрын
Help, my Bi-Pap Air Curve 10 just stopper communicating with the internet. I have a yellow circle with a line thru it where the bars were. I also have a blue flashing snowflake.
@cebruthius
@cebruthius 3 жыл бұрын
Finally Bilevel gets some love yay! But the DSX900 BiPAP AutoSV is the best BiPAP :P
@Freecpapadvice
@Freecpapadvice 3 жыл бұрын
Mail me one. Respironics never want's to send me any love. 🥲
@cebruthius
@cebruthius 3 жыл бұрын
@@Freecpapadvice I paid 1500 euros for mine :'(
@AsmageddonPrince
@AsmageddonPrince 3 жыл бұрын
@@cebruthius Would you like to pay 1500 euros for mine also? T_T
@catallergiccat
@catallergiccat 2 жыл бұрын
15:48 Hey Jason, do you mean increase the ipap to take care of the obstructive apnea?
@Freecpapadvice
@Freecpapadvice 2 жыл бұрын
No. Increase EPAP (lower of the two pressures) to eliminate obstructive apneas.
@randymeads3531
@randymeads3531 2 жыл бұрын
Excellent video!
@Winmx813
@Winmx813 3 жыл бұрын
I have the ability to buy this machine very very cheap I started cpap about 2 months ago and I just hate the relief from my 10 auto set …. Is this a good solution to make cpap more comfortable
@DeanoRolls
@DeanoRolls 3 жыл бұрын
I didn’t change any of the CPAP machine settings you talked about in this video (because I don’t have one) BUT all I know is I now comment from my channel like you told me to do. You say it...I do it!!!
@DeanoRolls
@DeanoRolls 3 жыл бұрын
I now fully expect your 40K subscribers will follow old deano’s walking in the woods videos!!!
@Freecpapadvice
@Freecpapadvice 3 жыл бұрын
"You say it...I do it!!!" - Deano 2021 "Deano, send me 3 million dollars!" - Lanky 2021 ....waiting....
@Freecpapadvice
@Freecpapadvice 3 жыл бұрын
If you could hike around while wearing a CPAP mask, holding a portable power pack in one hand (to power it), and a running CPAP machine in the other......no lie.....I will give you a shout out on my channel!!!
@DeanoRolls
@DeanoRolls 3 жыл бұрын
@@Freecpapadvice Uhhhhhhhhhh......... - Deano later in 2021
@DeanoRolls
@DeanoRolls 3 жыл бұрын
@@Freecpapadvice [adds item to Deano's to-do list]
@MrRazmataz
@MrRazmataz Жыл бұрын
Thanks for the "mini" class, a little more clarification on adjustments related to difficulties would be awesome. Example: my Aircurve 10 current settings are causing my exhalation effort to be too high. Would I just tweak the "EPAP" to find a better pressure? Thanks!
@VickysTuition
@VickysTuition 2 жыл бұрын
i think, @5:38 there is a mistake ! Higher Rise Time of 900ms means its a slower breath, NOT fast ! Similarly, Rise Time of 150ms very quickly changes from Epap to Ipap suggesting that it is a fast breath
@Freecpapadvice
@Freecpapadvice 2 жыл бұрын
If I said that wrong...apologies. Yes, longer time equals slower breath.
Respiratory Therapy - BiPAP vs. CPAP - How to adjust for ABGs?
20:58
Respiratory Coach
Рет қаралды 145 М.
Enceinte et en Bazard: Les Chroniques du Nettoyage ! 🚽✨
00:21
Two More French
Рет қаралды 42 МЛН
She made herself an ear of corn from his marmalade candies🌽🌽🌽
00:38
Valja & Maxim Family
Рет қаралды 18 МЛН
ResMed AirCurve 10 BiPAP Machine Review
5:38
RespShop
Рет қаралды 6 М.
How to Clean CPAP Machine, Tube and Mask | Tips for Cleaning
12:26
The CPAP Store
Рет қаралды 529 М.
AirCurve 11 Clinical/Provider Menu Settings Adjustment
8:05
CPAPmyway
Рет қаралды 4 М.
ResMed AirCurve 10 BiPAP Machine Overview
6:43
Sleeplay
Рет қаралды 43 М.
ResMed AirSense 10 for Beginners || Setup Guide
7:55
RespShop
Рет қаралды 26 М.
☠️Central Sleep Apnea is DEADLY - ResMed ASV Tutorial
14:39
CPAP Reviews
Рет қаралды 167 М.
How to Adjust Clinical Settings - AirSense 10
7:37
CPAPmyway
Рет қаралды 43 М.
Enceinte et en Bazard: Les Chroniques du Nettoyage ! 🚽✨
00:21
Two More French
Рет қаралды 42 МЛН